I never had troubles with the HV900's, and haven't had any that blew, so the following is based on what I've found useful troubleshooting/checking the regular Golden Reference HV. Often it can be helpful to work at lower voltages when trouble shooting. Paralleling the high resistor (is it 2x 450kohm in the HV900?) in the probe string to get the circuit in action at a lower output DC. If regulation doesn't work for some reason and things go bad it usually isn't really bad. Parallel the R8(2x